[Isocops] IBEX CEU SOH check for Orbits 163-167.
Mark Tapley
mtapley at swri.edu
Thu Oct 25 15:53:25 EDT 2012
IBEX instrument teams,
20 orbits behind. Well, it's been relatively busy. Starting
to catch up today.
--------------------
For Arc 163a: No obvious spin problems. Spin rate does show a
reaction to the eclipse around spin 9500. Hi Positive Collimator
current shows some "fuzz" around day 1.4. Otherwise nominal.
For Arc 163b: Spin error goes to large values near end of
orbit, although FractionValid doesn't drop below about 10% and
General Status plot looks OK. Otherwise nominal.
For Arc 164a: No obvious spin problems. IBEX-HI CEM back
count exceeds yellow limit a couple of times in the first day, not
near red limit; IBEX-Lo TOF Positive 5V Current monitor reflects
this, as does IBEX-Hi Total Counts. IBEX-Hi Positive Collimator
Current shows some strong "fuzz" about day 1.2. IBEX-Lo MCP Volt
monitor shows "fuzz" through day 2, reflected in Positive 12V CEU
Switch Current. Otherwise nominal.
For Arc 164b: No obvious spin problems, although as with last
orbit FractionValid drops near 10% near end of orbit. Otherwise
nominal.
For Arc 165a: Serious spin problems, starting day 0.7 or so,
with a curious drop-out in the data. ACS performance looks good, with
FractionValid dipping only to 50% or so near apogee. Otherwise
nominal.
For Arc 165b: Spin problems during day 2 and again at end of
orbit, reflected by FractionValid dipping to 10%, as well as near end
of orbit when it goes to ~0. Otherwise nominal.
For Arc 166a: No obvious spin problems. IBEX-Lo PAC Current
Monitor exceeds yellow limit at start-up for one sample. Otherwise
nominal.
For Arc 166b: No obvious spin problems, although
FractionValid goes to 0 near end of orbit. IBEX-Lo MCP Volt Monitor
shows 3 drop-outs. Otherwise nominal.
For Arc 167a: No obvious spin problems. IBEX-Hi CEM Back
Count Exceeds RED limit about day 1.25, for a noticeable period.
Yellow limit is exceeded for ~6 hours total. Back within limits by
day 1.5. IBEX-Hi Positive Collimator Current reflects this with small
current variations, as does the "noise" on IBEX-Lo LV Current and
Volt Monitors and IBEX-Lo TOF Positive 6V monitor, and to a small
extent by Positive 12V CEU Switch Current.
For Arc 167b: No obvious spin problems, although
FractionValid is dropping toward zero at the end of the orbit.
IBEX-Lo MCP Volt Montior has low "fuzz" days 5.5-6.5, with two
dropouts during that period; this is reflected by Positive 12V CEU
Switch Current as usual. IBEX-Lo PAC Current monitor exceeds yellow
limit at turn-on.
--------------------
...based on data at http://ibex.unh.edu/ibexhome/id-leops/status/ in
appropriate directories.
--
- Mark 210-379-4635
-----------------------------------------------------------------------
Large Asteroids headed toward planets
inhabited by beings that don't have
technology adequate to stop them:
Think of it as Evolution in Fast-Forward.
More information about the Isocops
mailing list